by
MikeD » Wed Nov 16, 2005 5:38 pm
Karine
Need a little more information. Why was it necessary to change the cold trap ? Is it the 3 mm inside diameter straight through, or low-flow narrow bore ? Did you pack it yourself ?
Assuming nothing else has changed and you are using the first type, the only thing I can think of is that it is possible for the trap material to move out of the heated zone if it is not retained firmly by the glass wool plugs. If you are using a low-flow trap this should not happen because there is a 1 mm restriction.
If the octadecane is 100% desorbed in the primary step it will be 100% adsorbed even on a weak sorbent like Tenax at say 30 C. It does not have to be cold. Toluene is the same. Because your toluene peak is OK that suggests you have no leaks. Therefore I believe your C18 is on the trap and the sorbent is not being heated to the temperature that the ATD is indicating, but heated to a temperature sufficient for toluene to come off.
Have you tried an n-alkane series to see where your apparent drop in recovery begins ?
If this does not explain your problem - sorry it's the best I can do. I have ignored many other possibilities because you say you have only changed the trap.