Impurity in glcyerine
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
UV spect show an abs at around 220nm. Is this a kind of impurity to be related to organic aromatic cmpd? If yes, how can I detect this using GC-MS. Can someone share experience. Plus, USP method for chromatographic system uses water as solvent. Will like to try to avoid that. Can someone advice other literature that I can refer to?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
The best source of information is the source of your glycerine.
Call the supplier and inform them of your findings. Ask them to explain the results and to inform you of the nature of the impurity.
If the problem is self caused you will save a lot of time. If the problem is vendor related then they will appreciate your informing them of the issue so they may correct it.
If it is a trace impurity and known to them, you will get good information and not hearsay. If not known to them they may help you determine the identity of the impurity quickly.
Good luck.
Call the supplier and inform them of your findings. Ask them to explain the results and to inform you of the nature of the impurity.
If the problem is self caused you will save a lot of time. If the problem is vendor related then they will appreciate your informing them of the issue so they may correct it.
If it is a trace impurity and known to them, you will get good information and not hearsay. If not known to them they may help you determine the identity of the impurity quickly.
Good luck.
-
- Posts: 6
- Joined: Sat Oct 29, 2005 3:58 am
Thanks Mr. Chromatography, we are the supplier. We are just curious to check whether one of our glycerine could have contained traces other organic compound with one of our newer GC-MS.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
I assume you do some wet chemistry tests to determine if aldehydes are present (or ketones or carboxylic acids or esters). These are the most likely contaminants with absorption of 220nm like you have detected with UV analysis. You can always try TLC to see if you can detect anything else. Have you derivitized glycerine to see if you can detect any other peaks by GC?
Sometimes using GC-MS is like using a microscope when you actually need wide angle binoculars.
Good luck.
Sometimes using GC-MS is like using a microscope when you actually need wide angle binoculars.
Good luck.
4 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science