Cleaned FID, reinstalled, now no signal on chromatogram

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
I have been using my HP 5890 for a few years now but am still a novice. I decided it would be a good idea to clean my FID, seeing as I never have, and am now running into problems. I will let you know what I did and what I have done to attempt to fix it.

I took it apart, cleaned the inside of the collector with a clean cotton swab, inspected the jet, which was actually pretty clean, and reassembled it. I started the machine and ran a standard and nothing showed up on my chromatogram. The flame was in fact lit (checked via condensation). I then took it apart again and was concerned that maybe I had damaged the jet (at this point I realized the removal tool in the GC drawer that I had used was not the correct one) and inspected it under the microscope. It appeared that the top had in fact been smashed a little.

I since purchased proper cleaning and proper FID removal tools and a replacement FID. I reinstalled everything but with the replacement FID, but when I ran it again, my chromatogram still read 0. And yes, the flame is once again lit. I am guessing the spring is not in the right position, but can't really tell what I am doing wrong. Also, at this point I am wondering if the spring is damaged. I am perplexed right now and am not sure exactly how to fix it.

I'm hoping this is an easy fix for all the experienced people out there. I am using a glass packed column for short chain VFA analysis (if it matters).

Thanks, Josie
Is electrometer turned ON [DET] [A] [ON] or [SIG] [A] [ON] ?
I think the detector is in fact on. I reassembled it again and used a stretched out paperclip to make sure the spring was in the groove. I am now getting a response but the same mM standard curve is coming out with less area than before.
Check if the jet is the correct one for packed columns.
Packed column jets have a wider orifice so the flame is different.
I purchased a replacement jet and reassembled the FID. The spring is very challenging to get into place and this time I used a paperclip to help me wiggle it into place. So far, things seem to be working. Thanks for the ideas!
5 posts Page 1 of 1

Who is online

In total there are 91 users online :: 0 registered, 0 hidden and 91 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 91 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry