I am working with a waters ACQUITY UPLC-TQD and am having issues with the consistency of the ratios between the quantifier and qualifier MRMs. The method I am working on is an opiates/opioids quant method that is analysing 41 different drugs, each drug has three transitions monitored (with the exception of two drugs).
The main problem that I am having is with 6-monoacetylmorphine (6-MAM) where the ratios of the quant transition and the two qualifier transitions are seemingly random and there is no consistency what-so-ever. Our tolerance is +- 20%.
The way that I have been measuring this is to do 5 injections each of a low, medium and high solution then taking the median and testing to make sure all of the ratios fall within 20% of this number. I am able to achieve acceptable results for most of the drugs in my assay by removing the results from the low solution but 6-MAM, morphine-3-glucuronide and morphine-6-glucuronide have issues with ion ratios at each of the three different concentrations.
I have tried removing 30 of the drugs from the assay to see if it is a case of asking the instrument to do more than it is capable of but this has resulted in very similar results.
Has anyone else had issues like this or know of a way to overcome these issues?
Any help would be greatly appreciated.
Thanks,
Jo