Page 1 of 1

Is the tailing factor acceptable????

Posted: Wed Nov 16, 2005 1:27 pm
by wxmwqr
I currently developed a new HPLC method, but the tailing factor is 1.3.
I tried to optimiz the mobile phase , colum, temperature and flow rate, but the tailing factor cannot to be improced.
Is the tailing factor acceptable?
I want to contibute the article to journal. Will the editor criticize it?

thanks a lot!

Posted: Wed Nov 16, 2005 6:54 pm
by tom jupille
I wouldn't worry about tailing until TF > 1.5. USFDA and ICH recommendations for system suitability set TF < 2. You're fine.

Posted: Wed Nov 16, 2005 7:24 pm
by MK
Tailing effects to integration precision, integration precision effects to peak area precision. Sharp symmetrical peaks gives consistent integrations. In my opinion, allmost any tailing (ok, up to 2 ;) is acceptable if the injection precision meets your criteria. However, you should show that the peak and the tail are pure and of the same component.

thanks

Posted: Thu Nov 17, 2005 2:49 am
by wxmwqr
thanks

Posted: Wed Nov 23, 2005 7:11 pm
by Chellie
<2 seems to be totally acceptable in anything i have read.
and i agree with MK, check to make sure that the tailing is actually just tailing and not a shouldering.

^_^