Choise of guard-column
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
Hi, I am using styrene/divinyl benzene, polymer PLPR-S 250X4.6 8um and I believe it is clogged. What is a good guard column for this column. We are separating peptides. Any suggestion appreciated.
-
- Posts: 103
- Joined: Sat Mar 12, 2011 4:02 pm
Its general recommendation: PLPR-S, 8um, 10-20mm X 4.6 or 4mm ID
You can also use only precolumn filter (PEEK or Ti) - for example, VICI Sure-Guard In-Line Filter (Ti, 0.5 or 2um).
P.S.: Also you must verify your beliefs
You can also use only precolumn filter (PEEK or Ti) - for example, VICI Sure-Guard In-Line Filter (Ti, 0.5 or 2um).
P.S.: Also you must verify your beliefs

2 posts
Page 1 of 1
Who is online
In total there are 329 users online :: 2 registered, 0 hidden and 327 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot] and 327 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot] and 327 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science