Baseline Problems

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
We are trying to validate a gradient method which uses a TFA buffer, A: 0.1% TFA and B: 0.1% TFA:ACN 20:80. We are using YMC C18 column. The diluent is 2% DMF in MeOH. The sample is difficult to dissolve.

The problem is that after a few injections a large hump is appearing before the main peak. It does not seem to increase or decrease with injection volumes, the increase/decrease appears to be random. There is no hump seen in the blank (both diluent and mobile phase). Any blank injection placed throughout the sample run is clean.

We have tried changing all the usual parameters, mp, column, system, sample prep but have not pinpointed the root cause.

Does anyone have any suggestions on where to go next?

:roll: :?

Assuming that you got some reasonable peaks, initially, (which should mean that your injection volume is low enough so that your solvent incombatibility doesn´t show), and assuming that you only injected clean standard it could be that air is causing the hump.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry