I am trying to find out what the industry standard is regarding check standards. More specifically, what calculation is performed and what are the limits that are set.
-
- Posts: 66
- Joined: Mon Jun 13, 2005 1:58 pm
Greetings,
I think your question is too open-ended.
What industry?
I am in the environmental testing biz, and our check standards are highly regulated, method specific, and show a wide range of acceptance criteria depending upon many factors.
For HPLC methods, quantifying at the ppb level, drinking water testing, we typically have to run checks every 8 hours, with deviation of +/- 20% vs. the initial calibration usually considered acceptable.
Evan Cooper
I think your question is too open-ended.
What industry?
I am in the environmental testing biz, and our check standards are highly regulated, method specific, and show a wide range of acceptance criteria depending upon many factors.
For HPLC methods, quantifying at the ppb level, drinking water testing, we typically have to run checks every 8 hours, with deviation of +/- 20% vs. the initial calibration usually considered acceptable.
Evan Cooper
-
- Posts: 6
- Joined: Fri Nov 04, 2005 3:21 pm
In the Pharmaceutical Industry, specifically the QC lab environment.
-
- Posts: 232
- Joined: Tue Sep 07, 2004 7:03 pm
The last GLP project I worked on, we used 15% error at several levels, except 20% error was allowed at the LLOQ. We used this for guidance:
http://www.fda.gov/cder/guidance/4252fnl.pdf
http://www.fda.gov/cder/guidance/4252fnl.pdf
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
And if it's cGMP, they will look for 1%.
The general rule can be summarized as: for the results to be valid, you have to demonstrate that you meet system suitability requirements. These will be specific to the particular method.
That said, the most conservative approach is to run a full system suitability before both and after running your samples, so that you can demonstrate that you were (presumably) OK in between. Intermediate check samples would be run in between as appropriate to avoid the possibility of failing system suitability at the end, thereby invalidating the whole set.
If you like to live on the edge, you could save time by running a check sample ("mini system suitability) at the beginnin to confirm retention, resolution, plates, etc. and then do the full system suitability (including repeatability) only at the end.
The general rule can be summarized as: for the results to be valid, you have to demonstrate that you meet system suitability requirements. These will be specific to the particular method.
That said, the most conservative approach is to run a full system suitability before both and after running your samples, so that you can demonstrate that you were (presumably) OK in between. Intermediate check samples would be run in between as appropriate to avoid the possibility of failing system suitability at the end, thereby invalidating the whole set.
If you like to live on the edge, you could save time by running a check sample ("mini system suitability) at the beginnin to confirm retention, resolution, plates, etc. and then do the full system suitability (including repeatability) only at the end.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 239
- Joined: Tue Nov 09, 2004 3:06 pm
Tom,
I disagree,
System suitability should be testet before analysis. Usually we do not have problems with resolution and plates. it just takes a number of injections until the system runs stable. This is especially true for for short isocratic methods.
Check samples are often run at the end. They are nice, yet i have never used (or needed) them.
Alex
I disagree,
System suitability should be testet before analysis. Usually we do not have problems with resolution and plates. it just takes a number of injections until the system runs stable. This is especially true for for short isocratic methods.
Check samples are often run at the end. They are nice, yet i have never used (or needed) them.
Alex
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Alex: I was making a joke with the last paragraph ("If you like to live on the edge . . ."). I certainly do not recommend that approach. Nonetheless, it is permissible.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 68
- Joined: Tue Nov 08, 2005 6:13 am
In my previous company, we have mentioned that sys suit once established will be applicable for next 24 hrs and after this 3 std injection shall be given. If sample ends before 24 hrs, one std injection will be given and rsd checked wrt initial injections.
In my present company, we r injecting a conrol std prep after 12 injection and after end of sequence. We r checking the RSD of control std with initial sys suit injections. It should not be more than 2.0%.
I have seen different practices in diff companies. USP has written one paragraph in the chromatography chpter about this. There are no specific criteria about it as per my knowledge.
Thanks
In my present company, we r injecting a conrol std prep after 12 injection and after end of sequence. We r checking the RSD of control std with initial sys suit injections. It should not be more than 2.0%.
I have seen different practices in diff companies. USP has written one paragraph in the chromatography chpter about this. There are no specific criteria about it as per my knowledge.
Thanks
jUST dO iT....
-
- Posts: 64
- Joined: Tue Oct 04, 2005 12:26 am
It is well known that bracketing system is to ensure that our system is doing well during the runs. Hence we need to use these checks not only at the beginning of the run but also in between and at the end of analysis.
In my lab the practice we follow is , if we use both standard and other systemsuitability parameters such as resolution ,plate count etc at the beginning of the run as systemsuitability then we have to use the check standard after every 3 samples and other parameters need to be checked at the end of analysis.
In my lab the practice we follow is , if we use both standard and other systemsuitability parameters such as resolution ,plate count etc at the beginning of the run as systemsuitability then we have to use the check standard after every 3 samples and other parameters need to be checked at the end of analysis.

-
- Posts: 6
- Joined: Fri Nov 04, 2005 3:21 pm
Thank you all for your responses.
10 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 25 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 25 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science