- Posts: 7
- Joined: Thu Aug 13, 2015 10:13 pm
I'm currently assisting a fellow faculty member at our university in trying to get her old Varian LC system back online after a year of being neglected. The setup is as follows:
- Varian Prostar 210 pumps (there are two of them)
- Varian 410 autosampler (although my friend says she does manual injections only)
- Varian Prostar 335 DAD
- Computer (Windows XP) running Galaxie software
The pumps and autosampler are connected to the computer through an RS-232 serial cable. The DAD is connected to the computer by an ethernet cable directly. There is no server - this is a standalone setup in a small lab.
Last year, the DAD refused to communicate with the computer and the system was shut down and neglected for a year. Recently, my colleague called in an Agilent person for $1500 who could not figure it out.
My colleague purchased another Prostar 335 (used off ebay) in case it was a problem with the hardware itself but she encountered the same problems. After the visit from the Agilent rep, the software was removed and reinstalled so basically I was starting from scratch when I worked on it.
Here's what I've done so far (from scratch).
- Galaxie software was cleanly installed before I arrived
- Installed Varian LC drivers from CD
- Ran Galaxie Configuration Manager Software
- Got the pumps responding again by setting up "communication engine configuration" for RS-232(PC) and following the prompts. I can now turn the pumps on and off using the software so I'm assuming its ok
- Running into problems connecting 335 detector to computer by ethernet cable. I've tried following the instructions in the manual for the driver as follows (please note I'm not a networking expert):
- Plugged ethernet cable into computer and detector. Windows network properties says "limited to no connectivity" yet gives me an IP address and subnet mask
- Used "Bootp" setup tool in Galaxie configuration manager
- using the autodetect feature it picks up the mac address of the 335 detector (most of the time anyway). A column appears for an IP address (defaulted to 0.0.0.0) - I enter a random value for this (since it's the not the same as my computer's IP address right?)
- To test the instruments, I set up a system in Galaxie configuration manager with the instruments that are attached to the computer. When I click on the DAD (335) tab, it comes up asking for an IP address but when I type in the IP address I gave the instrument in the bootp settings and click "check availability" - it keeps saying "not available" and I have no control over the detector.
I have tried fiddling with my computer's TCP/IP settings more times than I can count including manually assigning my computer a random IP address and so forth but nothing is working. I've also tried changing the ethernet cables and trying the second (backup) detector, no luck.
Any hints would be greatly appreciated!
TLDR: Can't connect a Varian Prostar 335 to a computer running galaxie using an ethernet cable (the machines won't talk to each other) despite trying many things as detailed above.