My coworker has been struggling with an issue on several 8260 volatiles instruments (6890/5973 with Enchon concentrator). The problem is Toluene-d8 (25 ppb) abundance, which should stay consistent for every run, drops out low with increased concentration of target analytes during a calibration, range 0.4-200 ppb. Toluene-d8 drops dramatically from target concentration 50-200. This also causes problems with linearity of toluene in the curve.
This problem is being seen across several instruments. I don't personally work with purge and trap systems and thought it could be an issue with the MS, something like saturation in the analyzer supressing signal for ions in toluene-d8, or EM too high. I'll list as many conditions as I have in front of me. Perhaps some setting is not quite right or someone here has experienced this before. Any thoughts would be greatly appreciated! Any additional info needed please let me know. Thank you.
purge vol = 5mL, purge temp = 35C, valve temp = 130C, line temp = 130C
trap type K (EST), purge ready = 35C, purge 11 min @ 35C, dry purge = 1 min @ 31C, desorb column pressure = 20 psi, desorb preheat 255C, desorb = 1 min @ 260C, bake 8 min @ 260C, purge flow 40 mL/min, trap back pressure 0 psi
6890N split, 2mm splitless liner, ratio 50:1, col flow 1.0 mL/min, 20.1 psi @ 45C
Oven initial temp 45C for 3 min, 15C/min to 230 for 3.3 min.
Column DB-624 20m, 0.18 mm, 1.0 um
MS 5973 BFB autotune EM=1565, foreline 219 mtorr, scan 35-260 amu, 5.98 scans/sec, transfer line 180C