Abnormal Retention time shifht-Can you help?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

11 posts Page 1 of 1
Hi

The compound I am working with has 3 OH group attached to P and a phenolic OH group as well.

For the analysis I am using a SB C18 column, Mobile Phase A is Water w 0.1% TFA and Mobile phase B is ACN w 0.1% TFA with a gradient elution.

The problem is during the qualification run, the RT for the main compound gradually shifted from 25.1 min to 19. 1 min in a span of 8 hrs continuous run. I dont think instrument is a problem here. I also dont think equilibration is a problem as well.

Sample prep is in EtOH/Water.

Can you offer any explanation? and what can I do to prevent it?
Thanks a lot.

Lilly

Generally speaking, it can be problematic when you are trying to re-equilibrate your column at 100% aqueous mobile phase. Maybe if you can start at 2% acetonitrile, it could be better...

I'll second Kostas's comment, but also ask what else is in the sample (i.e., what are the chances that you have irreversibly retained CRUD* building up on the column?

*CRUD = "Chromatographicall Retained Undesirable Debris"
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

First of all thanks for the replies.

My gradient starts from 30% organics, goes to 95% organics. So I dont think equilibration is a problem.

About the deposition in column; may be possible-- how can I avoid that?

Any other thoughts?

About the deposition in column; may be possible-- how can I avoid that?
Best way is not to let it get on the column in the first place (in other words, modify your sample cleanup). If that's not possible, you can use a guard cartridge to trap the crud (but you will have to change the cartridge periodically).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

how are your pressures? if it's dirt trapped in the column you should see some pressure increase. Is the temperature well controlled? are your mobiles premixed, if yes, can you loose some of the organic by evaporation? Is your compound stable or can it be changing to something else (can you look at the spectra of the shifted peak?)

I have been able to resolve the issue.

Thanks for all the help.

...and? How did you solve the problem?

curious in Reno...

Robert Haefele
Hamilton Company
rhaefele@hamiltoncompany.com

Just by increasing column equilibration time between runs.

Thanks all of you!

It was an equilibration problem after all :wink:
I think you would better disolve your sampl in mobile phase .it will decrease your equilibration time
looking for laboratory job in USA
11 posts Page 1 of 1

Who is online

In total there are 27 users online :: 3 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry