- Posts: 17
- Joined: Fri Nov 11, 2005 3:18 pm
The compound I am working with has 3 OH group attached to P and a phenolic OH group as well.
For the analysis I am using a SB C18 column, Mobile Phase A is Water w 0.1% TFA and Mobile phase B is ACN w 0.1% TFA with a gradient elution.
The problem is during the qualification run, the RT for the main compound gradually shifted from 25.1 min to 19. 1 min in a span of 8 hrs continuous run. I dont think instrument is a problem here. I also dont think equilibration is a problem as well.
Sample prep is in EtOH/Water.
Can you offer any explanation? and what can I do to prevent it?
Thanks a lot.
Lilly