Sensitivity Issue with Refractive Index Detector
Posted: Thu Mar 03, 2016 5:52 pm
Hello everyone,
I am developing a method for sugar analysis (sucrose, glucose etc) using Agilent 1100 series refractive index detector using a TSKgel Amide-80 column. I can see my sugar peaks but they are very small even when I am injecting 2% wt/wt solution.
My conditions are as follows:
Equilibration time = 2 hours
RID heater temp = 35 C
purge time = 20 min
flow rate = 1mL/min
Injection volume = 5, 10 or 20 uL
Mobile phase = ACN/H20 80:20 in same bottle
Column temp = 40 C ( I started out with 50 C. Results comparable to 40 C)
Also I see a huge peak in the beginning (~3.4 minutes for a 30 min run), even in blanks. This peak intensity is around 400,000 nRIU whereas those of the compounds of interest are around 40,000 units.
I would appreciate any suggestions as to how to increase sensitivity.
Thank you.
I am developing a method for sugar analysis (sucrose, glucose etc) using Agilent 1100 series refractive index detector using a TSKgel Amide-80 column. I can see my sugar peaks but they are very small even when I am injecting 2% wt/wt solution.
My conditions are as follows:
Equilibration time = 2 hours
RID heater temp = 35 C
purge time = 20 min
flow rate = 1mL/min
Injection volume = 5, 10 or 20 uL
Mobile phase = ACN/H20 80:20 in same bottle
Column temp = 40 C ( I started out with 50 C. Results comparable to 40 C)
Also I see a huge peak in the beginning (~3.4 minutes for a 30 min run), even in blanks. This peak intensity is around 400,000 nRIU whereas those of the compounds of interest are around 40,000 units.
I would appreciate any suggestions as to how to increase sensitivity.
Thank you.