FAME question
Posted: Tue Mar 01, 2016 2:06 pm
by erdno
Dear all!
I have to analyse fatty acids (c12-c20) in soil and i have problem how to prepare my sample. I would extract it with hexane then add potassium hydroxid in methanol to derivatize them. Then i would analyse it with GC-MS.
If anybody could help me with any suggest, or perhaps methods or literature i would be really thankfull.
Thank u very much!
Erdno
Re: FAME question
Posted: Tue Mar 01, 2016 6:52 pm
by DR
Search the forum for FAME methods, we've beaten them to death over the years. Not sure about the soil matrix effects, but I suspect you'll find BF3/MeOH to be a more popular derivitizing agent...
Re: FAME question
Posted: Tue Mar 01, 2016 6:55 pm
by GOM
I agree
Regards
Ralph
Re: FAME question
Posted: Tue Mar 01, 2016 8:39 pm
by Peter Apps
Are you interested in only the free fatty acids, or do you want the fatty acid profile of the lipid content of the soil ?
Re: FAME question
Posted: Thu Mar 03, 2016 2:07 pm
by Consumer Products Guy
Yes, that's the real question, I've done thousands of FAME assays. If just looking for composition of free fatty acids or their salts, simply extract with methanol and heat 5 minutes after adding some methanol-H2SO4 or BF3-methanol, then cool and extract into hexane, adding NaCl solution to help separate the layers.
If you want to include triglycerides composition, I'd saponify with KOH-methanol, then add my methanol-H2SO4 or BF3-methanol, then cool and extract into hexane, adding NaCl solution to help separate the layers.
Re: FAME question
Posted: Fri Mar 04, 2016 2:36 pm
by MSCHemist
For simple free fatty acids I do chloroformate esterfication. I use isobutyl chloroformate and isobutanol because I can just barely get isobutyl acetate separated from the catalyst for the reaction pyridine. I quench the reaction with 1N NaHCO3 to remove excess chloroformate.
For total FAMES I like the mild methanolysis procedure here
http://www.ncbi.nlm.nih.gov/pmc/articles/PMC2817593/
You can make the reagents out of common lab chemicals like aqueous 35%HCl. It gets diluted in the final reaction mix to 1.2% HCl and 2.4% H20 which is insignificant. I very occasionally get asked for a basic fatty acid profile.
I can also to lactic, malic, tartaric, citric, succinic, and fumaric using a deriviative of the above method though I replace the methanol with isobutanol and heat it at 90 deg for 2 hours then add 1ml of 1N NaHCO3 to quench the HCl.
Re: FAME question
Posted: Fri Mar 04, 2016 5:18 pm
by GOM
In addition to the above helpful comments and observations I have found that the following website by W W Christie (The Oily Press) to be an excellent summary of lipids, their form, hydrolysis, derivatization methods and mass spectra
http://lipidlibrary.aocs.org/Analysis/c ... mber=19210
Regards
Ralph