What could be the problem? Problem with injector port?
Possible troubleshooting.
Division of Pharmaceutics
Central Drug Research Institute
Lucknow, India
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Carry over residual peaks. With manual injector flush your syringe properly (sufficient number of times). Have proper time gaps between injections. See to it your column is not overloaded. Try this, keep injecting blank and observe the peak area. Is their a decrease, ideally it should decrease and at some point it will vanish. Is your analyte highly hydrophobic? if this is the case flush with strong solvent.I am analyzing a compund (with literature reported method). When I inject blank solvent after standard solution or sample run, a small peak of the analyte compound (same UV spectrum-using PDA detector) appears repeatedly run to run at same RT with variable peak areas.
What could be the problem? Problem with injector port?
Possible troubleshooting.