Page 1 of 1

Instrument for Corrosive Gas Analysis

Posted: Fri Nov 04, 2005 4:40 pm
by jaj
Does anyone have any suggestions as to a GC system that can be used to analyze corrosive(fluorinated)gases?

corrosive gases GC

Posted: Wed Nov 30, 2005 10:50 pm
by chromatographer1
I will assume you wish to use an FID.

Parts that contact the heated sample or the flame: ie, electrodes, etc should all be gold plated.

Gas Lines should be thick wall Stainless steel glass lined, or deactivated with a silica deactivation at a minimum.

I would ask Valco to suggest the best metal for the valve and sample loops, Monel 400 might be best or perhaps Hastelloy C-22.