Method parameters:
agilent 7890
Diluent-Dimethylacetamide
Column-75m x 530um x 3.0um db-624
flow- 13ml/min nitrogen
splitless
column temp prog- 35c for 15min-25c/min to 150c for 11min
inlet-140c
det-220c
Advertisement
Discussions about GC and other "gas phase" separation techniques.
I have the idea that if you provided details about how you are introducing the samples - headspace vs liquid injections for instance - you might get more answers.Does anyone have any ideas about solving this problem?
Is it done with a 6-port valve injection, an automated syringe injection, or a balance pressure injection - in other words what hardware are you using ???This is a headspace method:
Incubation temp. is 120 c
injection volume is 800ul
I agree that the inlet temp. is low,
but this is a validated method that i recived from a manufacturer.