muGC » Wed Feb 25, 2015 4:10 pm
HPLIS is not required for LPG analysis. PLIS alone is more than adequate to vapourize the sample providing the concentration of C1 is below 1% for single phase injection.
With PLIS there is no pulse heating applied to the sample as there is adequate thermal energy derived from the Injection port to gasify the sample.
PLIS has been successfully used in LPG analysis with superior results to Rotary valve injection as the transfer zone is extremely short (less than 80 uL) with no interruption of carrier gas flow.
Agreed, I use Agilent's version of the PLIS (they call it the HPLI). It works great and provides superior results to the traditional LSV configuration. It's not a carrier gas flow issue though, it is the pressure drop between the column head and the LSV. This pressure drop allows a phase transition from liquid to gas and cause band broadening. Comparisons of PLIS and traditional LSV injections show the PLIS with sharper peaks due to a nearly 100% liquid injection. However, if you purchase a GC configured for ASTM D2163 from Agilent, PAC, etc., it will just be configured with a traditional LSV and GSV.
Traditional LSV configuration and vapor sampling are still acceptable.