Zero counts in positive ESI, good negative mode signal
Posted: Thu Feb 18, 2016 11:34 pm
Hi everyone,
I've never seen this problem before. I'm using an Agilent 6130 multimode MSD coupled to a 1200 Series HPLC. All of a sudden, I am seeing no signal whatsoever in positive mode, not even background ions. I'm talking ZERO counts of any ions with a totally flat TIC. Negative mode background looks typical, with random noise fluctuations in the baseline (~15,000 counts) and strong signal when an analyte elutes and is introduced into the detector.
I also get no observable ionization in positive mode even when using the autotuning program.
My current spray chamber settings have been used successfully with this method many times and are:
MultiMode dual ESI/APCI
Scan range: 100-1000 amu
Fragmentor: 70, Gain 1.00, Threshold 150, Step size 0.10
Drying gas flow: 5 L/min
Nebulizer: 40 psig
Drying Gas Temp: 350 C
Vaporizer Temp: 200 C
Capillary Voltage: 2000 V both positive/negative
Corona Curent: 5 µA both positive and negative
Flow rate: 1 ml/min
The only recent change to the instrument that I'm aware of is that we've recently installed EasyAccess walk up software. I'm worried I may have a serious electronics problem in the MSD. I've tried resetting the board but without any luck. Has anyone come across anything like this before?
I appreciate the help. Thank you!
DJA
I've never seen this problem before. I'm using an Agilent 6130 multimode MSD coupled to a 1200 Series HPLC. All of a sudden, I am seeing no signal whatsoever in positive mode, not even background ions. I'm talking ZERO counts of any ions with a totally flat TIC. Negative mode background looks typical, with random noise fluctuations in the baseline (~15,000 counts) and strong signal when an analyte elutes and is introduced into the detector.
I also get no observable ionization in positive mode even when using the autotuning program.
My current spray chamber settings have been used successfully with this method many times and are:
MultiMode dual ESI/APCI
Scan range: 100-1000 amu
Fragmentor: 70, Gain 1.00, Threshold 150, Step size 0.10
Drying gas flow: 5 L/min
Nebulizer: 40 psig
Drying Gas Temp: 350 C
Vaporizer Temp: 200 C
Capillary Voltage: 2000 V both positive/negative
Corona Curent: 5 µA both positive and negative
Flow rate: 1 ml/min
The only recent change to the instrument that I'm aware of is that we've recently installed EasyAccess walk up software. I'm worried I may have a serious electronics problem in the MSD. I've tried resetting the board but without any luck. Has anyone come across anything like this before?
I appreciate the help. Thank you!
DJA