Advertisement

Help with hydrocodone bitartrate tailing issue

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi all - I'm trying to improve on a method that i previously developed for the analysis of hydrocodone bitartrate via HPLC.

The problem is that the hydrocodone peak has a tailing factor of ~ 0.85 due to a long slow baseline drift occuring right before the bulk of the drug elutes.

I've tried two different systems thus far with no positive results
1. RP18 column with with KH2PO4 (pH 6 & 4) and NH4H2PO4 (pH 3) buffer
2. C18 column with KH2PO4 (pH 6 & 4) and NH4H2PO4 (pH 3) buffer

the pH of the solution is ~3 and the column is not being overloaded (ie: the baseline drift can be detected at very low concentrations) I have a hunch that the column is still to retentative or i'm just totally overlooking something very important.

any insight would be greatly appreciated.

HB tails (our limit is 1.5 at 5% peak height).
We run a ~40 min. phosphate/MeOH gradient (pH3) on a C-8 column at fairly high temperature. HB elutes at abou 13 min.
Thanks,
DR
Image

Your tailing is due to residual silanol interaction. There are bunch of columns on the market with extensive end-capping, you can also try to lower pH to suppress silanol activity. We are using a different approach. Check the following link for similar compound (dextromethorphan) with the symmetry of 1.0-1.05. The ligand in Primesep B2 column masks any residual silanol:

http://www.sielc.com/application_066.html

What is good about this approach is that you can retain your anionic counter ion too.

Here is another one for basic compound:

http://www.sielc.com/application_065.html

0.85 means fronting, not tailing, right? What's your injection volume, injection solvent, and column i.d.?
4 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry