desolvation gas and cone gas quattro lcms
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
Cone gas is on but desolvation gas is off. Nitrogen gas is set to the right pressure. do you know how I can fix this?
-
- Posts: 10
- Joined: Fri Aug 08, 2014 1:13 pm
Hi,
press API button (nitrogen gas) and COL (argon gas)! Then on the gas flow tab you can manage to adjust the flow (L/hr).
Cheers
press API button (nitrogen gas) and COL (argon gas)! Then on the gas flow tab you can manage to adjust the flow (L/hr).
Cheers