External Standard Method

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi all,
In USP <621> CHROMATOGRAPHY:
External Standard Method: The concentration of the component(s) quantified is determined by comparing the response(s) obtained with the sample solution to the response(s) obtained with a standard solution.

Is there any specific statement in USP or other official documents require the concentration of the standard solution to be similar to that of the sample solution?

Thank you all in advance.
Not that I'm aware of. It's common practice since it can allow you to eyeball the approximate concentration of your sample. This can give you a leg up on beginning to troubleshoot an issue if the peaks aren't similar. That said, as long as your results are demonstrated to be accurate, it doesn't really matter what your standard concentration is.
Generally if you run only one concentration of standard than you should use approximately the same concentration as your sample. The pickle is to know you b value from standard curve.

If the b value in equation y= ax + b is significant different from 0, than running different concentration from your sample will make you results underestimated or overestimated. In this case making the same concetration as your sample is right way.

If you know that you b value from standards curve is not significant different from 0, that propably using different concetration from your sample should make your result accurate.

Kreall
I will always prefer to generate a standard curve, covering the region I find relevant.
Two rationales for that. 1: If you only have one level standardization the other/lower end of the curve will be assumed to be zero and that is very, very seldom the case. So, that was the first source of determination error. 2: How do you know what concentration the sample would have, especially if you test stability samples or recovery samples etc.? So, that was the other source which might result in necessity for re-analysis if the std. and sample concentrations are too far from eachother.
Actually, there’s a third rationale for generating std. curve: You test the linearity in every analytical series, which is a typical SST parameter.

Best Regards
Learn Innovate and Share

Dancho Dikov
4 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry