Running CSH column with 100% aqueous?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Can we run a Waters C18 CSH column with a 100% aqueous mobile phase without having any problem?

If not, what should be the minimum methanol or acetonitrile % in the mobile phase?

Thanks in advance.
AFAIK, should be Ok with 100% water.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Unless your samples and standards are in 100% water you're not buying yourself anything with 100% mobile phase. I'd tend to run at least 5% organic phase and worry a lot more about the final solvent system of the sample.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA