carryover with peptide analysis
Posted: Wed Oct 26, 2005 8:25 pm
Hello all. I am developing a method to analyze peptides with molecular weights in the 3000 - 4000 dalton range, in a biological fluid by LC-MS/MS. I have determined some useful MRM transitions using synthetic standards of the peptides, but have hit a stumbling block with the LC. I've tried a few different columns, pH modifiers, and tried ACN versus MeOH as organic modifier. I get good chromatography from a Luna C18(2), 2.1x50mm, 3µm, 100 angstrom pore size column, with water-ACN gradient, and 0.1% formic acid as pH modifier. The problem is with carryover, and something that looks like adsorption.
When I run a solvent based calibration curve, I see carryover as high as 10% (depending on the peptide) with the blank following the high standard. A non-injection blank after the high standard indicates that some carryover appears to be coming from the column, although I suppose I can't rule out active sites elsewhere in the flow path. A flow-injection experiment with standards and blanks indicates that some carryover is coming from the autosampler. My needle-wash solvent is presently 1:1 MeOH/water, which works well with some small molecules but might not be ideal here. Any suggestions for wash solvents or sample additives that will help minimize carryover will be appreciated. I have already tried adding bovine serum albumin (at 10x the molar concentration of my high standard), and it actually seems to worsen the carryover.
Also, any suggestion as to a column that won't hang on to my compounds through to the next analysis, will be appreciated. So far, all the columns I have tried have been silica based, C8 or C18. For mobile phase modifiers, I'm somewhat limited to what gives good MS sensitivity. I've already ruled-out acetic acid, formic acid/ammonium formate, and high pH (ammonia) for these peptides.
Another bad thing I have noticed: the signal seems to drop of rapidly when comparing the high standard to lower standards (or increases rapidly from low to high, depending on your point of view). In other words, let's say I get 100,000 area counts with my high standard. I expect to have something close to 10,000 area counts at 1/10 the concentration, and 1,000 area counts at 1/100 the concentration, but actually I might see 2,000 counts at 1/10 the concentration, and not detected at 1/100! I wonder if this could be caused by adsorption, and if so, could it be related to the carryover described above. I am using plastic (no glass) for all standard preparation and storage.
If you can't tell, analysis of biomolecules is new to me. I don't recall ever having this much difficulty with small molecules!
When I run a solvent based calibration curve, I see carryover as high as 10% (depending on the peptide) with the blank following the high standard. A non-injection blank after the high standard indicates that some carryover appears to be coming from the column, although I suppose I can't rule out active sites elsewhere in the flow path. A flow-injection experiment with standards and blanks indicates that some carryover is coming from the autosampler. My needle-wash solvent is presently 1:1 MeOH/water, which works well with some small molecules but might not be ideal here. Any suggestions for wash solvents or sample additives that will help minimize carryover will be appreciated. I have already tried adding bovine serum albumin (at 10x the molar concentration of my high standard), and it actually seems to worsen the carryover.
Also, any suggestion as to a column that won't hang on to my compounds through to the next analysis, will be appreciated. So far, all the columns I have tried have been silica based, C8 or C18. For mobile phase modifiers, I'm somewhat limited to what gives good MS sensitivity. I've already ruled-out acetic acid, formic acid/ammonium formate, and high pH (ammonia) for these peptides.
Another bad thing I have noticed: the signal seems to drop of rapidly when comparing the high standard to lower standards (or increases rapidly from low to high, depending on your point of view). In other words, let's say I get 100,000 area counts with my high standard. I expect to have something close to 10,000 area counts at 1/10 the concentration, and 1,000 area counts at 1/100 the concentration, but actually I might see 2,000 counts at 1/10 the concentration, and not detected at 1/100! I wonder if this could be caused by adsorption, and if so, could it be related to the carryover described above. I am using plastic (no glass) for all standard preparation and storage.
If you can't tell, analysis of biomolecules is new to me. I don't recall ever having this much difficulty with small molecules!