Endrin/Pesticide issues on Agilent 6890A
Posted: Tue Feb 09, 2016 5:49 pm
Hi all,
I have been reading this forum the last few weeks after stumbling on it accidentally and I must say it is very informative. So, I figured it's about time I joined! I have been working on GCs for the past 4-5 years, Varians and Agilents, specializing in PCBs and Pesticides mostly various methods. I have some background in DRO, Herbicides, and ETU as well. So, on to my issue..
We recently started using an Agilent 6890A for 8081 analysis. On the Agilent, we are looking decent but we have a serious problem with Endrin Aldehyde. The curve is consistent, and passes well against 2nd source CCV on the alternate pest GC (a Varian) but on the Agilent Endrin Aldehyde recovers some 25% higher than other analytes immediately following curve.
One would suspect breakdown, but when running the breakdown mix, breakdown of Endrin is practically non existent (<1%). The recovery in the CCV doesn't seem to get better with time/injections which suggests it is not something baking off that is quenching/adding response. And this is regardless of whether we just clipped column/installed new inlet or ran 10 curves back to back to back. This GC has not yet seen samples as we are still in the initial QC stage.
My initial thought was that something in the liner/inlet is quenching the response of Endrin Aldehyde that then bakes off before the CCV runs (only 7 injections later) resulting in a "high" response. But this doesn't make sense as you would think that after that point, re-calibrating would be spot on due to whatever "gunk" present baking off and no longer affecting response. Unfortunately, this is not the case.
The only differences between the two GCs are the column types and that the Varian has silco-steel injectors while the Agilent (problem) does not. Any ideas?
Thank you for the help and suggestions.
MCA
I have been reading this forum the last few weeks after stumbling on it accidentally and I must say it is very informative. So, I figured it's about time I joined! I have been working on GCs for the past 4-5 years, Varians and Agilents, specializing in PCBs and Pesticides mostly various methods. I have some background in DRO, Herbicides, and ETU as well. So, on to my issue..
We recently started using an Agilent 6890A for 8081 analysis. On the Agilent, we are looking decent but we have a serious problem with Endrin Aldehyde. The curve is consistent, and passes well against 2nd source CCV on the alternate pest GC (a Varian) but on the Agilent Endrin Aldehyde recovers some 25% higher than other analytes immediately following curve.
One would suspect breakdown, but when running the breakdown mix, breakdown of Endrin is practically non existent (<1%). The recovery in the CCV doesn't seem to get better with time/injections which suggests it is not something baking off that is quenching/adding response. And this is regardless of whether we just clipped column/installed new inlet or ran 10 curves back to back to back. This GC has not yet seen samples as we are still in the initial QC stage.
My initial thought was that something in the liner/inlet is quenching the response of Endrin Aldehyde that then bakes off before the CCV runs (only 7 injections later) resulting in a "high" response. But this doesn't make sense as you would think that after that point, re-calibrating would be spot on due to whatever "gunk" present baking off and no longer affecting response. Unfortunately, this is not the case.
The only differences between the two GCs are the column types and that the Varian has silco-steel injectors while the Agilent (problem) does not. Any ideas?
Thank you for the help and suggestions.