Hi,

I am working on stage 2 content uniformity test where the first 10 samples were analyzed in one sample set and the next 20 samples were analyzed in another. Does Empower 2 have the capability to process both of these sample sets as one so that it would calculate AV value?

We currently have custom calculations built to calculate AV, Average, and Standard Deviation.
It would have been better if for you to have added the next 20 samples to the sample set of the first 10 that failed.

Aside from the above, I do not believe there is a way to do what you want in Empower 2.

In Empower 3 you could simply create a processing only sample set method that is a combination of your n=10 and n=20 runs.