Peak Purity and S/N in FLD detector

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Does anyone know how to check peak purity and s/n ratio in FLD detector?

I am using Agilent HPLC with OpenLab Chemstation software
For peak purity, you would have to compare excitation and/or emission spectra at various times across the peak. I'm not aware of standard software that does that.

S/N can always be done the old-fashioned way: print out the chromatogram and meazure the signal and the noise.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
For FLD spectra, If I remember correctly, could be wrong, you have to capture the sample in FLD cell and stop flow, then you press left mouse on detector and chose something like take sample spectrum. You have to have installed 3D spectra acquisition license.

For calculating S/N, you go under Data Analysis, report, system suitability, edit noise time and chose the time frame of your noise. Specify the report as Performance + Noise
3 posts Page 1 of 1

Who is online

In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry