Nebuliser gas flow remains OFF

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi guys,

I'm having a problem with the nebuliser gas flow in a Waters Quattro LC mass spectrometer.

Before, when I turned on the gas, both nebuliser and desolvation gases went up togheter, but now the nebuliser gas remains off. The gas pressure is ok and the desolvation gas flow is right. I opened fully the nebuliser flow control valve, but nothing changes.
Does anyone have an idea about what might be happening?

I appreciate any help!

Karina
Hi,

Problem solved!
The technitian explained that the "nebuliser" indication in the screen in fact means "cone gas" in the Quatto LC. Not very intuitive. :?

Then, I adjusted the cone gas valve and the nebuliser gas turned up in the screen.
Very simple, but it took me many hours...
Maybe this post can help someone newbie like me in the future. :)
thanks for posting the solution. You're right - it is really helpful for people searching in the future
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry