Glutaraldehyde - DNPH derivatization
Posted: Wed Jan 27, 2016 6:44 pm
by JI2002
Hi All,
I'm working on analysis of glutaraldehyde by HPLC following derivatization with DNPH (containing H3PO4). The problem I have is that multiple peaks with two major ones were observed to be related to glutaraldehyde after derivatization. LCMS analysis of the derivatized glutaraldehyde standard showed the two major peaks are isomers and both are the expected derivative. Searching online showed no report of two peaks for glutaraldehyde-DNPH derivatization by HPLC. Has anyone experienced the same problem?
Thanks.
Re: Glutaraldehyde - DNPH derivatization
Posted: Wed Jan 27, 2016 7:39 pm
by Consumer Products Guy
We've assayed glutaraldehyde by HPLC after DNPH derivatization and only obtained a single peak.
Posted: Wed Jan 27, 2016 7:52 pm
by JI2002
Thanks CPG for your reply. I'm puzzled by the two peaks. The derivatization procedure is simple: 0.8 mL 5 mg/mL 2,4-DNPH (acidified with 1.5% H3PO4) was added to 2 mL of glutaraldehyde standard in ACN, incubate @40C for 1 hour, centrifuge and inject. Column XDB C8, 150X4.6, 3.5 um. mobile phase A 5% ACN in water mobile phase B 100% ACN
45/55 A%/B% for 3 minutes to 100 B% at 10 minutes, UV@360 nm.
Posted: Thu Jan 28, 2016 8:50 am
by bunnahabhain
Is your second peak DNPH?
Posted: Thu Jan 28, 2016 1:40 pm
by JI2002
Thanks, bunnahabhain. No, the second peak is not DNPH. DNPH elutes before the derivative peaks. My guess is that the glutaraldehyde reference standard is not pure. I've requested COA from the glutaraldehyde vendor. At the same time, I'm ordering a glutaraldehyde-DNPH standard in order to figure out the cause of the two peaks.
Posted: Thu Jan 28, 2016 2:44 pm
by bunnahabhain
If you have 1,3-Cyclohexanedione in your lab, try this procedure:
http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3934741/
I tried for C13 and C17 and it worked.
Posted: Tue Dec 13, 2016 8:23 pm
by bundy5555
It is normal. Depending on what chromatographic conditions you use. We just integrate two peaks together as one peak to get the area.
Posted: Thu Dec 15, 2016 8:34 pm
by JMB
Off the top of my head, don't you expect two peaks ?
R1-CH=N-R2 with H and R2 syn
and
R1-CH=N-R2 with H and R2 anti
with the same at the other end of the molecule; because of symmetry you only have two different stereoisomers.
Posted: Fri Dec 16, 2016 3:37 pm
by JMB
Yes, I forgot the one SYN one ANTI configuration.
Posted: Fri Dec 16, 2016 7:40 pm
by JI2002
cis and trans configuration isomers are the most likely reason for the double peaks. Thanks.
Posted: Wed Jan 25, 2017 2:47 am
by mattmullaney
Agreed--you've got your answer already...this Uchiyama guy has written a ton about DNPH derivatization of asymmetric aldehydes:
"Isomerization of aldehyde-2,4-dinitrophenylhydrazone derivatives and validation of high-performance liquid chromatographic analysis", Journal of Chromatography A, Volume 996, Issues 1–2, 9 May 2003, Pages 95-102, by Shigehisa Uchiyama, Masanori Ando and Shohei Aoyagi.
this is one example. Not much you can do about it...EPA TO-11A is fraught with this kind of nonsense...controlling the pH of the derivatization of both standards and samples helps a wee bit (also including a lag time for the reaction to come to equilibrium), but always there are two isomers of the shorter-chain aldehydes.