Currently performing a linearity study on a related substances HPLC-UV method for a non-UV absorbing molecule. The method uses Nitric acid as mobile phase and a PRP-X100 column; the analyte peaks elute as negative peaks.

A linearity study was performed on concentrations ranging from LOQ (0.1% of spl conc) to 120% of limit concentration (0.28% of sample concentration); this gave a good correlation co-efficient value (> 0.99) however the y-intercept wrt 100% spiked solution obtained was -17%.

A seperate linearity study performed at higher concentrations gave a y-intercept of 0.5%.

It is evident that in this case, the lower the concentrations, the more negative the y-intercept being obtained.


Would appreciate if anyone can give me any pointers with regards to:

- what is the significance of such a low -ve % y-intercept value (-17%)?

- how does this result impact the suitability of the method?

- are there any suggested modifications that can be done to the method to improve the y-intercept? The sample concentration cannot be altered since it is the highest possible concentration to give an adequate S/N ratio at LOQ concentration, while maintaining an adequate specificity between the main peak and other known impurities in the test solution.
The problem is not the y-intercept, but the x- intercept - the non-zero quantity of analyte that gives a zero peak area,and correspondingly increased LOD and LOQ.

Just a guess - are you sure that there is nothing (besides the mobile phase) that absorbs UV eluting at the same time as your analyte ?

Peter
Peter Apps
also, how linear is your calibration curve? R-squared values are often a dreadful measure of linearity - they can look really high, while the residuals (deviations from the straight line) are obviously very non-random (high at the ends and low in the middle, or vice versa). If present, deviation from linearity can be either a genuine effect of the detection system (in which case I personally think it better to find a fit that matches the detection system, rather than pretend it's linear, but I don't have to deal with regulatory bodies who can't understand anything that doesn't go in straight lines) - or it can be a problem such as Peter described, or a limited number of binding-sites for an analyte in effect titrating out the analyte at low concentration.
You fail to mention your wavelength that you are using. Nitrate absorbs intensely in the low UV; you may be seeing the effects of trying to measure small changes in a large absorbance.

As lmh points out, definitely run residuals on your curve. We normally toss relatively linear curves into log space and then do a least squares fit; it does a better job of fitting all of the points, not just the high end of the curve.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
4 posts Page 1 of 1

Who is online

In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry