Seen lab results that when diluted by a factor of 10, it raises the LOQ by a factor of 10 making it an unusable number. Isn't the LOQ a measure of instrument performance and not take into account dilution?
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
The original is probably not a true LOQ which is why I prefer the statistical calculation of the LOQ based on the sum of the residuals (standard deviation). Factors that influence the LOQ are injection volume, detector, flow, interferences...
If you normally see 10ng/ul as your LOQ, then if no prep is involved the LOQ of the sample would be 10ng/ul. If you dilute the sample ten fold, you still see an LOQ of 10ng/ul on the instrument, but that has now become 10ng/0.1ul of sample or 100ng/ul. The LOQ of the instrument doesn't change when you dilute, but you can no longer have the same sensitivity versus the actual sample.The original is probably not a true LOQ which is why I prefer the statistical calculation of the LOQ based on the sum of the residuals (standard deviation). Factors that influence the LOQ are injection volume, detector, flow, interferences...
Right, so if I have my LOQ calculated from LOQ = Blank response + 10*STD Dev
Of a set of blank samples then that value should not change when I dilute a sample, right?