Chemstation Peak Identification if Outside RT Window

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
Hello, I'm a new participant here. I have lots of GC and IC experience and have just started working with HPLC systems.

In TotalChrom and Chromeleon if a peak in a single data file is slightly outside the retention time window it is easy to assign it as one of the calibrated analytes in the method and subsequently integrate and report without updating the method.

I'm using a system with OpenLab Chemstation software and I don't seem to be able to assign an identity to a peak if it falls outside an RT window.

I've enlarged the method RT windows slightly to rectify the problem but I'm curious if it can it be done as part of reprocessing or recalculating?

Thanks,

Mike
The default calibration window from RT+/-5%. Increase it to 10% will solve the problem for most of the case.
2 posts Page 1 of 1

Who is online

In total there are 38 users online :: 0 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry