Changing Elution Order of Peaks on FID
Posted: Tue Jan 19, 2016 5:30 pm
Hi, all;
Early last December I posted a problem with late-eluting peaks switching order on a GC we were setting up. We finally resolved that by swapping injectors as a last resort.
I am setting up another 7890 GC, and am seeing the same swap in elution order of the last six peaks. The configuration is simple, really - Hydrogen carrier, Split/Splitless injector, HP-1 20m x 0.18mm ID column, into an FID. Gas hydrocarbon sample is introduced into the injector via a 6-port Gas Sampling Valve. An easy analysis I've set up/run a hundred times.
I get great, fast chromatography for all compounds at the front/middle of the chromatogram. However, by Ethylbenzene, sensitivity drops to very poor, m/p-xylene looks like Bart Simpson, and the last several peaks elute in the wrong order.
This is the "bad" chromatogram:

this is what it looks like on the eight working GCs:

The net amount of sample on column is approximately the same for both chromatograms; I've tried fiddling with the split ratio on the new system, but it doesn't increase my back-end sensitivity at all, just overwhelms the Methane/Ethane and causes them to co-elute.
I tried different liners, glass wool, no glass wool, re-installing column into injector, tightening the injector weldment nut, leak checking (ok), flow checking (ok), and even swapped out the injector/EPC entire unit, since that worked last time
. Nothing has made things better (maybe even worse sensitivity with the new SSI).
I've ordered another column from a different manufacturer, but not very hopeful that will make much difference. Has anyone else ever run into this strange problem? Any ideas what else I can try to resolve this?
I'd really appreciate any suggestions, I can't think of much else to try.
Early last December I posted a problem with late-eluting peaks switching order on a GC we were setting up. We finally resolved that by swapping injectors as a last resort.
I am setting up another 7890 GC, and am seeing the same swap in elution order of the last six peaks. The configuration is simple, really - Hydrogen carrier, Split/Splitless injector, HP-1 20m x 0.18mm ID column, into an FID. Gas hydrocarbon sample is introduced into the injector via a 6-port Gas Sampling Valve. An easy analysis I've set up/run a hundred times.
I get great, fast chromatography for all compounds at the front/middle of the chromatogram. However, by Ethylbenzene, sensitivity drops to very poor, m/p-xylene looks like Bart Simpson, and the last several peaks elute in the wrong order.
This is the "bad" chromatogram:

this is what it looks like on the eight working GCs:

The net amount of sample on column is approximately the same for both chromatograms; I've tried fiddling with the split ratio on the new system, but it doesn't increase my back-end sensitivity at all, just overwhelms the Methane/Ethane and causes them to co-elute.
I tried different liners, glass wool, no glass wool, re-installing column into injector, tightening the injector weldment nut, leak checking (ok), flow checking (ok), and even swapped out the injector/EPC entire unit, since that worked last time
I've ordered another column from a different manufacturer, but not very hopeful that will make much difference. Has anyone else ever run into this strange problem? Any ideas what else I can try to resolve this?
I'd really appreciate any suggestions, I can't think of much else to try.