Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
I am suffering in the extra peaks brought from phosphate buffer in my HPLC analysis. the reagents used for the preparation are both ACS grade not HPLC grade which are much much more expensive. does anyone know the difference between these two grades? will it help if I replace with HPLC grade?
-
- Posts: 430
- Joined: Thu Nov 12, 2009 3:28 pm
ACS-grade phosphate contains some heavy metals that HPLC-grade does not. These accumulate on some columns during equilibration and then elute in well-defined peaks during the gradient. You can get rid of them by preparing a solution of the ACS-grade reagent and passing it through a column packed with a low-pressure mixed-bed ion-exchange resin. Alternatively, put the resin beads in a fine mesh bag and swirl it through the solution. There's a detailed procedure for this by Mant and Hodges in Methods in Enzymology (@ 1996 or so). If you want the exact reference, I can cite it on Monday when I get back to the lab.
Having said that, it might be easier and perhaps cheaper just to buy HPLC-grade phosphate reagent.
Having said that, it might be easier and perhaps cheaper just to buy HPLC-grade phosphate reagent.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
It’s far from certain the phosphate salt is the culprit here. It might as well be the ACN and indeed the water.
I’ve seen all the variations of the above and in addition to that contaminations stemming from the mobile phase containers, filters (if the eluents are filtered) etc.
My point here is the need for identifying the source of the trouble and address it properly.
Best Regards
I’ve seen all the variations of the above and in addition to that contaminations stemming from the mobile phase containers, filters (if the eluents are filtered) etc.
My point here is the need for identifying the source of the trouble and address it properly.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 715
- Joined: Wed Aug 04, 2010 2:45 pm
I fully agree with Andy.
My experience is that even USP standards are not 100% pure and HPLC grade buffers can be qualified against ACS and USP standards. That saves some money and makes your life easier.
My experience is that even USP standards are not 100% pure and HPLC grade buffers can be qualified against ACS and USP standards. That saves some money and makes your life easier.
Gerhard Kratz, Kratz_Gerhard@web.de
4 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Adsense [Bot] and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Adsense [Bot] and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science