I've an issue with Gas chromatography Thermo series Trace 2001. I can’t run my samples because the GC gives an error that the split flow isn’t reaching. The split flow should be set 93 ml/min. but the split flow can reach up to 80 ml/min. The leak check and the K- value test passed.
Here are the current conditions:
Detector:260°C
Injector: 230°C
Carrier gas: He
Split flow: 93 ml/min.
Column Pressure : 0.55 [bar]
Oven: initial Temp. 195 [°C], heat up to 220°C (2,5 °C/min.), end temp. 220°C (hold 12 min.)
My instrument is limited to a split flow of 200ml/min and this analysis ran 100 times successfully with this condition as mentioned above.
A leak in or around the inlet? I didn't believe, because the leak check passed.
Is a blockage in the split vent trap? If yes, how can I check it?
Thanks for your support.
Dao