Old column bleeds on transfer to new inlet

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hello
I wonder if anyone has experience with an old column that bleeds extensively after change of inlet? I am using a column TG5-MS that has been previously used for PyGCMS analyses, consecutively 4 years, up to a T ramp of 210degC and occasional bake-out till 250degC. This column has a max T of 330degC/350degC. After changing to the next available inlet, performed column conditioning and purging, as well as an overnight bakeout, the bleeding remains high, such that my hexadecane injection standard of 140ppm remains undetected. Do you have a solution on what I can try? Thank you very much.
Lynn
Welcome to the forum.

Pyrolysis HC often puts heavy muck onto the column that subsequently bleeds off. If you can tell which was the inlet end of the column when it was used for pyrolysis, then cut 1 m off that end and re-install. If you cannot tell which was the inlet end, then cut 1 m off both ends.

Peter
Peter Apps
Thank you so much, peter. I will go try it.
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry