Lost intensity - pesticide

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi all,

I'm devolping a method to analyse 6 pesticides, using the following conditions:

HPLC- Waters 2695
MS - quattro premier XE
Mobile phase; A= UpH20 + 0.1% Acetic acid/ B = Methanol + 0.1% Acetic acid

MRM - 2 transitions

I was running a 10ug/L standard to check repeatability, with good results and then after +/-10 runs, I started to lose intensity on one compound (Metaldehyde).
I already prepared the standard several times, cleaned the cone, check capillary for leaks, change column, change solvents, re-tune several times with great narrow peaks, but still having the same problem.
The other pesticides (quinmerac, carbetamide, pendimethalin, metazachlor, propyzamide) are fine with great repeatability, good peak shape and intensity.

Any ideas?

Regards
Are you injecting from the same vial over and over again?

Try preparing your standard then separating it into several vials or even one vial per injection.
Problem solved! Thanks richiekichi!
3 posts Page 1 of 1

Who is online

In total there are 24 users online :: 3 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google Adsense [Bot], Semrush [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry