UPLC Gradient
Posted: Wed Dec 16, 2015 12:43 pm
by smartin
I would need some advice on how to construct a gradient, please.
I am working with and UPLC-UV.
I have started with a gradient (A:B) of (95:5) to (0:100) for 20min at a flow of 0.3 mL/min. My peaks of interests are at 3.4 and 5 min.
How should I proced, to reduce the time run?
I mean, maybe one option is (A:B) of (95:5) to (70:30) for 10min at a flow of 0.3 mL/min?
Otherwise, I think I should increase the flow, at least at 0.5 mL/min. Do you agree?
Please, all help will be welcome!!!
Thanks a lot!!
Re: UPLC Gradient
Posted: Wed Dec 16, 2015 9:12 pm
by Hollow
both could be possible even together.
if there aren't any peaks after 5 min then, you can just trim your gradient after the peak came out, but keep the slope of your gradient (%/min) the same. Otherwise you may change the separation
So in your case you could trim your gradient from 5% to 33.5%B in 6 min
If there are some unwanted compounds after 5 min, keep your gradient slope the same until all peaks of interest are out then fast increase the %B to 100% to flush everything out. eg. 0/5%, 6/33.5%, 6.2/100%
Second you can increase the flow rate as much as your system and column can handle, but then further reduce the gradient step to keep the gradient volume (time gradient x flow) the same.
So with 6 min x 0.3 ml/min = 1.8 ml
-> with 0.5 ml/min, your new gradient would be 5% to 33.5%B in 3.6 min
This should keep your separation constant if there aren't any other effects from dwell volume or pressure issues
Re: UPLC Gradient
Posted: Fri Dec 18, 2015 9:10 am
by smartin
Hi!!
Thank you very much for this great explanation!!
It has worked!! :-)
Thanks again,
Re: UPLC Gradient
Posted: Sun Dec 20, 2015 8:50 pm
by Hollow
Welcome :D
Thanks for feedback