New GC configuration for biogas analysis
Posted: Wed Dec 16, 2015 2:52 am
Greeting guys,
I'm a postgraduate student from Malaysia. My supervisor is planning to get a new GC set for our lab. Below are the specifications:
Shimadzu Gas Chromatograph Model GC-2014AT
complete with:
- 2 x DINJ-2014 (dual packed injector)
- 2 x TCD-2014 (Thermal Conductivity Detector)
Technical Specifications
Column Oven
- Temperature range: Room temperature +4 oC - 400 oC
- Programmed rate setting range: -250 to 250 oC/min
Injector
- Independently temperature controlled injector units with AFC
Detectors:
Thermal Conductivity Detector (TCD)
- Temperature range: up to 400oC
- Dynamic range: 10^5.
Display:
- 240 x 320-dot graphics display (30 characters x 16 lines)
P/N: 221-70027-34
Gas Sampling Valve
Gas sampler with 6 port valve with adapter, MGS-4
P/N: 221-42574-02; 221-42868-91
Stainless Steel Packed column
Shincarbon-ST 50/80 (6m x 3mm ID)
For your information, our lab is interested to quantify biogas composition in % (v/v)
Based on our previous results the gas composition: hydrogen (25-45%), nitrogen (about 0-20%), oxygen (0-20%), methane(in the absence of hydrogen gas, about 35-55%), carbon dioxide (35%-45%).
I would like to have your comments/opinion whether this GC configuration is sufficient for our analysis? Thanks in advance.
I'm a postgraduate student from Malaysia. My supervisor is planning to get a new GC set for our lab. Below are the specifications:
Shimadzu Gas Chromatograph Model GC-2014AT
complete with:
- 2 x DINJ-2014 (dual packed injector)
- 2 x TCD-2014 (Thermal Conductivity Detector)
Technical Specifications
Column Oven
- Temperature range: Room temperature +4 oC - 400 oC
- Programmed rate setting range: -250 to 250 oC/min
Injector
- Independently temperature controlled injector units with AFC
Detectors:
Thermal Conductivity Detector (TCD)
- Temperature range: up to 400oC
- Dynamic range: 10^5.
Display:
- 240 x 320-dot graphics display (30 characters x 16 lines)
P/N: 221-70027-34
Gas Sampling Valve
Gas sampler with 6 port valve with adapter, MGS-4
P/N: 221-42574-02; 221-42868-91
Stainless Steel Packed column
Shincarbon-ST 50/80 (6m x 3mm ID)
For your information, our lab is interested to quantify biogas composition in % (v/v)
Based on our previous results the gas composition: hydrogen (25-45%), nitrogen (about 0-20%), oxygen (0-20%), methane(in the absence of hydrogen gas, about 35-55%), carbon dioxide (35%-45%).
I would like to have your comments/opinion whether this GC configuration is sufficient for our analysis? Thanks in advance.