How to Integrate 35900E signal in Chemstation C.01.06
Posted: Mon Dec 14, 2015 8:12 am
by polychiharu
Hi all,
I have existing 1200 chemstation instrument and is connected to the 35900E Series II ADC.
There is no problem to interface both instrument to the PC. I can see the online plot from Chemstation C.01.06.
The problem that I faced now is couldn't capture the signal in the windows when i run the sequence.
The window is like when we run a sequence, the signal will be captured and we can integrate the data. However, it couldn't capture the 35900E signal.
Kindly advise is there any setting that I missed out?
Re: How to Integrate 35900E signal in Chemstation C.01.06
Posted: Mon Dec 14, 2015 3:32 pm
by Consumer Products Guy
I've run across this exact issue before: signal shows in Online Plot but is not saved in the run or sequence (no vertical red line at injection in online plot).
You need a remote cable from the 35900 box to the stack, like on ours the 35900 remote cable connects to remote on the UV detector. Got it?
Posted: Wed Dec 16, 2015 2:52 am
by polychiharu
Thanks for the advice. It is work once connected the remote cable from 35900E to the 1200 stack. It really made my day. Thanks for the sharing.
Posted: Wed Dec 16, 2015 1:20 pm
by KM-USA
Winner, winner, chicken dinner !
Part of being smart is remembering and learning from my past mistakes...
Posted: Thu Dec 17, 2015 3:30 pm
Is he going to E-mail that chicken dinner to you?