Hi Trishia,
I would recommend a much cheaper option.
There are Special "two hole ferrules" available, where you can install two columns in one MS.
I am shure you will found this at different suppliers.
greetings
But this needs two columns, and although I've never tried it I'm pretty sure that you can't fit two columns through the transfer line tip. And even if you could, unless you play around with pressures and flows it increases the total flow to the MS and makes it harder for the pumps to pull a vacuum.
Peter
0.18 ID columns have an outer diameter near 0.4mm and a 0.53ID column has an outer diameter of 0.8mm so theoretically you could slide two 0.18ID columns through the transfer line (at least of an Agilent) at the same time. The problem would be that the two holes in the two hole ferrule would be wider than 0.8mm and I would worry that at the point where they meet the interface they would be crushed when the ferrule is tightened. Now if the interface was beveled it might work, but that would require altering the interface and possibly preventing the ferrule from sealing. The hole in the nut would also need to be enlarged to accept the two columns.
This setup would be far superior to the one I am running with the Tee ( I run an Rxi624SilMS and a Rxi5SilMS to do volatiles and semivolatiles on the same instrument). Flow is easy to control with inlet pressure switching.
It is an interesting idea, and I would jump in it if the interface is compatible. Maybe I will do some research on this in the future.