I have a serious problem with the GC-MS. It is Thermo Finnigan Trace GC coupled with Polaris GCQ plus.
Last year we upgraded Xcalibur 1.2 to Xcalibur 1.4. Last Friday I tried to open tune window and there has been a message saying that control of the system has been lost would you like to restart? Also the Ion Source Temperature was 0 C.
I pressed ok and everything worked ok and Ion source reached again the temperature set point (200 C).
Last Monday when i opened Tune there was the same message again, but this time when i pressed ok to restart the system there was this message: "System Initialization Error! we had no control at all.
What is worst is that we found the chip that regulates communication between the MS and data system destroyed and the Transfer Line Heater and Temperature Sensor burnt.
The problem is we cannot where the problem starts. Is it the chip that is burnt (and why is this happening?) and then all the other problems follow. Or is the loss of communication between the data system and the ms that creates this or the Transfer Line. The main board is new so as the transfer line.
Thank you!