We are running Thermo trace 1310 GC-FID. We couldnt get benzene peak. Could you please recommend GC condition. We used NIOSH 1501 method and no success.
thank you very much in advance,
Regards,
Mungun
Advertisement
Discussions about GC and other "gas phase" separation techniques.
It has a rather small retention index coming off at 650 or so on db-1 column (same area as hexane so the solvent peak (what solvent are you using) could cover it up. A thicker film would be better.
642 DB-1
648 OV-1
663 CP-Sil-8CB
936 Supelcowax-10™
938 DB-Wax
Liquid injection or headspace?
I get excellent retention and peak shape for benzene on a Rtx-624 (Restek). The one I use most is the 60 m x 250 µm x 1.4 µm version. It's a thicker film (MSChemist's suggestion) so if you have late eluters, it might be problematic. My conditions:
Inlet: 240 °C, helium carrier gas at 15.5 psig
Oven: 40 °C for 2 min. 40-240 °C at 10 °C/min., 240 °C for 18 min.
Detector: FID, 30:300:25 H2:Air:helium makeup, 240 °C
99% of the time, I'm sampling the headspace. Usually solid-phase microextraction.