Advertisement

Strange peak on FID and PFPD

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi everybody,

I work on a GC-17A. My system is equipped with two lines:
- one connected to a FID, with a Restek ShinCarbon ST 80/100 column
- one connected to a PFPD, with a Agilent GS-GasPro column
I use the first line for analyzing methane, and the second line for sulphur compounds. I work on natural gases.

Everything seems to work fine, but on my both channel, I have a very strange peak (the first one), and I can't find what it is... Maybe someone can help me ?
I put a chromatogram for example. It is FID response. For PFPD it is a negative peak.

Image

Thank you for any help
My guess would be that it is the disturbance peak from the injection, technically the balance gas. If correct the balance gas has more conduction in the FID than your carrier for FID and less for PFPD. As it is there in excess it is good that the the peak is very small and not surprising it has a tail.
Great ! Thank you for the reply !
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry