Agilent 5975B high water and nitrogen content
Posted: Tue Dec 01, 2015 12:24 pm
Hi @ all,
we have some trouble with a high water and air content in our system. We did a leak check with Argon and 1,1,1,2-Tetrafluoroethane but there is no leak detectable. Our water content today was around 20%. The same for Nitrogen. We are using a special software for our system and not the Chemstation. One problem is that our pressure gauge seems to be broken because the values sometimes jump from 5x10^-6 to 3x10^-3 Torr in Milliseconds. Now we have a flow of 2.8mL/min and a pressure of 1.16x10^-5 Torr (The Agilent Guide says it should be at least 1.86x10^-5 Torr). If we can trust the pressure gauge this value would be ok.
Picture 1 shows the abundances from last Friday. The system is only flushed with Helium as carrier gas. So shouldn't Helium have the highest Abundance if we don't find any leak? Because Nitrogen and Oxygen are much higher.
1)

Picture 2 shows the abundances from today. Yesterday we putted the gas bottle out of our gas bottle cabinet to connect it with a 2m long tubing directly to the GC and heated it out over night (we thought that the Air/Water might come from the lines in the wall). When tho oven was opened the water signal became some oscillations.
2)

Picture 3 shows the tuning results from the last days. The air-water parameters on Friday 151127 1456 seemed good. The tuning parameters before changing was 151130 0800. Actually we increased the system temperatures (MS source 230°C, quad 210°C, oven 200°C and interface 200°C, the default temperatures before the run were MS source 150°C, quad 130°C, oven 40°C and interface 200°C) to clean the system overnight. But the air-water parameters seemed bad when we did the tuning test this morning and we didn't detected any leak.
3)

All in all we have these poor results since we cleaned the source and exchanged both filaments but we are quite sure that we did everything correct.
We hope that anyone can give us some hints and help us. Thank you very much.
The Medusa-Team
we have some trouble with a high water and air content in our system. We did a leak check with Argon and 1,1,1,2-Tetrafluoroethane but there is no leak detectable. Our water content today was around 20%. The same for Nitrogen. We are using a special software for our system and not the Chemstation. One problem is that our pressure gauge seems to be broken because the values sometimes jump from 5x10^-6 to 3x10^-3 Torr in Milliseconds. Now we have a flow of 2.8mL/min and a pressure of 1.16x10^-5 Torr (The Agilent Guide says it should be at least 1.86x10^-5 Torr). If we can trust the pressure gauge this value would be ok.
Picture 1 shows the abundances from last Friday. The system is only flushed with Helium as carrier gas. So shouldn't Helium have the highest Abundance if we don't find any leak? Because Nitrogen and Oxygen are much higher.
1)

Picture 2 shows the abundances from today. Yesterday we putted the gas bottle out of our gas bottle cabinet to connect it with a 2m long tubing directly to the GC and heated it out over night (we thought that the Air/Water might come from the lines in the wall). When tho oven was opened the water signal became some oscillations.
2)

Picture 3 shows the tuning results from the last days. The air-water parameters on Friday 151127 1456 seemed good. The tuning parameters before changing was 151130 0800. Actually we increased the system temperatures (MS source 230°C, quad 210°C, oven 200°C and interface 200°C, the default temperatures before the run were MS source 150°C, quad 130°C, oven 40°C and interface 200°C) to clean the system overnight. But the air-water parameters seemed bad when we did the tuning test this morning and we didn't detected any leak.
3)

All in all we have these poor results since we cleaned the source and exchanged both filaments but we are quite sure that we did everything correct.
We hope that anyone can give us some hints and help us. Thank you very much.
The Medusa-Team