by CE Instruments » Tue Sep 13, 2005 3:56 pm
There are many UK customers using Hydrogen carrier as it is the most efficient carrier gas and less expensive than Helium. If used with care you should have no problems. Newer GCs with DPFC (Digital Pressure and flow control) will switch off in the event of a major leak which is an added safety feature. In 15 years of GC I have only heard of one Hydrogen "accident". A user of a Carlo Erba 4000 series (circa 1970) set their GC up and went for coffee. On their return they were puzzled to see all their colleagues hiding under benches ! They had installed the column without ferrules, not done a leak check , just lit the FID and left. Once the GC filled up with Hydrogen (head pressure flow control) it ignited. Fortunately the spring loaded door held (design feature, not guarenteed) and no-one was harmed. The engineer was asked to visit to write off the GC for insurance purposes. Unfortunately the GC in questions was fine , if a little circular rather than square and the GC

and was put back into service, still with Hydrogen carrier.