Area percent on TCD calculation on overloaded column

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I inherited a method which I am obligated to try and use. Viewing the main peak I can see it is not gaussian but has front edge leading peak indication column overload or method condition is not optimal. My question is this... will it make a calculation difference if a peak is front edge leading peak in an area percent method vs. a more symmetrical peak shape?? I always thought that really no matter the shape the amount coming out the TCD was the same, and the integration still would find the same area.
If you're overloaded to the point of fronting peaks, then you're probably not in the linear response range for that analyte anymore. This means your values will be artificially low. Dilute your sample by half and by 10x.

What's your injection volume?
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry