Hi LC gurus,

May I get your advice on an baseline issue that I've never had with this Agilent 1260 LC? Even without actually doing a run, if I switch the mobile phase with mixed water and acetonitrile to 100% acetonitrile, after about 2 min once the pressure reached the lowest, the baseline would all of sudden increased and get stabilized. Then once I switch it back to the mixture of water and acetonitrile, again after 2 min once the pressure reached the highest, the baseline sharply decreased and stabilized there. I tried to change to fresh mobile phases, changed from 100% acetonitrile to 50% ACN/50% Methanol, changed to a different column. Still had the exactly same problem. Your suggestions would be greatly appreciated!
It is the problem that you changed the viscosity of your Mobil phase composition, no problem with your instrument and column!
Gerhard Kratz, Kratz_Gerhard@web.de
2 posts Page 1 of 1

Who is online

In total there are 38 users online :: 0 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry