I have a question about Method Detection Limit determinations in GC for hydrocarbon analysis, using GCFID & GCTCD. We prepare an MDL standard at 1.5-5 X the assumed MDL, run ten replicates, and then quantitate the resulting signal using our normally calibrated instrument.
My question is, does the calculated MDL standard concentration you get matter? For example, if your spike level is 1.7 ppm for a 0.5 ppm MDL, does it matter whether you get 0.9ppm, 1.2ppm, or 2.3ppm consistently for the calculated concentrations of ten replicates? Is it reasonable to accept a calculated value that is lower than the expected concentration, or a bit higher, if they pass the required statistical calculations for the MDL?
One employer insists that it matters, but a former stated that it does not. Can’t find a definitive answer on the web.
I'm worried and starting to doubt myself! Anyone care to share an opinion so I can have a little peace of mind?