Sample prep: hydrazine derivatization for LC-MS/MS analysis
Posted: Wed Nov 18, 2015 5:20 pm
Hi everyone,
Synthetic chemist, doing some proper analytical work here. So any tips/pointers is appreciated.
I'm trying to develop/optimize a quantitative method for analysis of diketones in plasma samples which require an excess of hydrazine to form the corresponding hydrazones for detection. For the reaction to go to completion, a large excess of hydrazine (aq) is used. However with those concentrations present some issues may present on the instrument (Waters Xevo TQ-S)... In the literature there seem to be some different approaches; evaporation of the sample mixture, which hopefully will remove the hydrazine, followed by re-dissolution in an appropriate solvent. Or simply just inject with the excess hydrazine present...
I'm a bit skeptical about injecting samples containing large amounts of hydrazine (1-3 mM), but is evaporation (via N2-stream, combined with slight heating) of the samples (sometimes followed by re-dissolution in MeOH and evaporation again) enough?
My current plan is basically:
1. To plasma, add internal standards (structurally similar diketones) and hydrazine solution.
2. Add CH3CN/MeOH to precipitate proteins.
3. Incubate to complete derivatization.
4. Centrifuge, remove aliquot for analysis.
5. Evaporate to dryness.
6. Re-dissolve in mobile phase.
7. Inject.
Thoughts?
7. Inject.
Thoughts?