Calculation of LOD when there is caryover in the blanks

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
the s/n of the peak in the blank was subtracted from the s/n of the peak in the 0.01% std injection
Is this the typical way to calculate LOD if a peak is observed in the blank injection or should the peak height of the blank have been subtracted from the peak height of the 0.01% std before calculating the s/n?
Welcome to the forum.
the s/n of the peak in the blank was subtracted from the s/n of the peak in the 0.01% std injection
Is this the typical way to calculate LOD if a peak is observed in the blank injection or should the peak height of the blank have been subtracted from the peak height of the 0.01% std before calculating the s/n?
The peak in the blank is noise. If you are going with the usual optimistic S:N > 3 criterion for LOD, then the peak from the standard must be > 3 x (blank peak + baseline noise).

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry